Adjusting the Monochromator Entrance Slits

The mono entrance slits are set using the micrometers on the upstream side of the mono tank. The one on the left is for the vertical slits, the one on the right is for the horizontal.

attachment:slits.jpg

Vertical slits

The vertical slit assembly consists of four slots of decreasing heights. The slit heights and their corresponding micrometer positions are:

height

approximate position

3.0 mm

1.830

2.0 mm

1.520

1.0 mm

1.105

normal use

0.5 mm

0.725

use for high resolution XANES

When you change energies, it is a good idea to optimize the entrance slits by watching the signal on the feedback detector I00.

Horizontal slits

The horizontal slit assembly consists of a series of slots on increasing width in a fork hanging down in front of the beam. The vertical extent of each slot is quite tall compared to the beam, so positioning of the mike is not so important.

width

approximate position

9 mm

0.80

18 mm

1.05

normal use

25 mm

1.30


Community/X23A2/MonoSlits (last edited 2009-10-09 19:51:15 by localhost)