X23A2 Beamline Evaluation: Detector Noise
| Slideshow ^ |< << Slide 4 of 6 >> >| |
This is a pretty simple test -- just a series of time scans in which the detectors are polled repeated for 3 minutes while dwelling under fixed conditions. To make the latter parts of this test, I borrowed a 50 micron tungsten foil from the EXAFS equipment pool. At 10 keV, the energy of this test, this foil is about 9.4 absorption lengths. The ion chambers were filled with nitrogen and the harmonic rejection mirror was in place at an angle of 0.2 deg.
The W foil was affixed to the edge of the sample stage with masking tape in such an orientation that it would be simple to block the beam with either the horizontal or vertical edge simply by moving the sample stage. This means that these noise tests are made by comparing the It and Ir chambers rather than I0 and It. All three chambers are of the same type and have been in service for the same period of time. Of course, the I0 chamber is 10 cm shorter than the other two, but I assert this test is applicable to any measurement at our beamline. It was much less work to set up the test this way than to move ion chambers around.
Each of these scans shows the ratio of the signals on the It and Ir chambers as a function of dwell time.
- First test, 10 keV, full beam
- Second test, no beam
- Third test, 10 keV, beam half blocked by a horizontal knife edge
- Fourth test, 10 keV, beam half blocked by a vertical knife edge
The noise is slightly higher in the half-blocked scans. I suspect that the culprit is the silly job I did mounting the foil. The mechanical instability (floppiness) of the foil is probably the cause of the higher noise level in those two scans. In any case, the noise level in our ion chambers and signal chain is excellent.
| /100_Calibration /200_Harmonics /300_Linearity /400_Noise /500_Quality /600_Resolution |